Test Strip Design and Reading
A project log for Universal Glucometer
Did your test strips just get more expensive? Has the drug store run out of strips for YOUR meter? Like to be able to use any test strip?
tom-meehanTom Meehan 07/17/2016 at 04:460 Comments
Overview of the some of the strip design details as well as the methods used in reading data from the strips and using it to determine blood glucose.
The Amperometric Method of determining blood glucose measures the electric current generated at a specific point in time by the glucose reaction (with either Glucose Oxidase or Glucose Dehydrogenase enzymes). This method, primarily uses either the:
Self Biased Configuration (2 electrode) Transimpedance Amplifier
or:
Counter Configuration (3 electrodes) Transimpedance Amplifier
Illustration taken from:
Both of these methods amplify the current produced by the reaction of glucose on the test strip and convert it to a voltage. The voltage is then fed into an Analog to Digital Converter and the digital signal is interpreted by the microcontroller, using a regression equation, to output a blood glucose concentration.
Below are some of the different electrode configurations in the reaction area (at the tip of the strip) used by different manufacturers.
Figure 4. Small volume electrochemical blood glucose monitoring strips with their top cover layer removed.
  • W: Working Electrode
  • C/R: Counter/Reference Electrodes
  • F: Fill Detrection Electrode
  1. (A) BD test strip, with electrodes in recessed wells in an insulating layer.
  2. (B) Accuchek Aviva, its fill electrodes doubling as hematocrit-compensation electrodes.
  3. (C) Ascensia Contour.
  4. (D) Precision Xtra with electrodes covered by a mesh.
Figure 4. taken from: