2 Years
A project log for Precision Voltage and Current Reference
A little something to check your digital multimeter accuracy occasionally.
bud-bennettBud Bennett 06/29/2021 at 02:070 Comments
Digikey Voltage Box: (2.5hr soak, 64F ambient, 100PLC, HighZ input)
   Reading              Scale            Other                                1 yr Drift (Vref + DMM)
100.00097mV      100mV           Null + Avg 20 samples       1.57 uV
0.9999591V         1V                  Null + Avg 20 samples       -51 uV
2.499842V           10V               Avg 20 samples                  82 uV
4.999121V             10V               Avg 20 samples                  -279 uV
10.000018V          10V               Avg 20 samples                  -12 uV
AliExpress Box: (6hr soak, 62F ambient, 100PLC, HighZ input)
99.99982mV       100mV           Null + Avg 20 Samples       -1.99 uV
1.0000619V            1V                Null + Avg 20 Samples       50.9 uV
2.500164V            10V               Avg  20 Samples                 24 uV
10.000056V          10V               Avg 20 Samples                  26 uV
Current Reference:
64F
    Value               Scale                                                  1 yr Drift (Iref + DMM)
100.00127uA       100uA     Avg 20 samples                 8.14 nA
0.9999867mA     1mA        Avg 20 samples                 9.7 nA
9.999337mA        10mA     Avg 20 samples                 -2.137 uA
Conclusions:
After 2 years the meter and references are well within the expected tolerances for both the voltage and current ranges.