DAV5 V3.01 Raman Spectrometer
The only thing worth doing, is the thing worth doing right!
This is my place to play...My own little private Idaho!
UPDATE 01/12/2018 6:31:PM
Tactical flashlight scan @ 2ms CCD 8-Bit (TCD1304DG) detector circuit test:
A better 1st mirror alignment and a 2nd 150mW laser line test:
The plot above has a FWHM value of 1.005nm resolution.
Fixed the 1st mirror position, I have much better resolution now :)
Another test scan today using my Aries 150mW green laser and a 0.1mm slit width @ 3ms integration time:
A little bit of clipping can be seen even though I had the variable polarizing lens at 75% still got a resolution factor of 2nm :)
New test scans using the 8-Bit CCD detector (TCD1304DG)
Final version of the Motor Control Targeting Interface unit for the diffraction grating turret control system:
Excel chart data:
New CCD scans using a 32mm achromatic coated lens @ the 1st collimation mirror and a compact fluorescent lamp, below is the raw data of various scans:
Below is the processed data of my CFL lamp:
and finally, the FWHM data from the above spectrum:
A calibrated Solux lamp spectrum (mine,) 4700K/50W/12vdc/1.2A/0.10mm slit width
Turret @ zero position (baseline) 0.00 nm/ Integration time: 150ms
2400ln/mm diffraction grating:
7 Dark Count scans I did to determine which integration time is the most ideal and it appears to be 50ms, 15ms and 10ms:
CCD detector Allmon & Haffner v1.1 is done.
-Testing Platform
PLX-DAQ v2.11 Dual Trace Transfer to Excel
New 8-Bit TCD1304 CCD detector driver board
New schematic
8-Bit Main SCH 1
8-Bit Sheet 2
Cost analysis comparing Raman Spectrometers and Lasers from StellerNet Inc. to the DAV5 V3.01 3D printable Raman Spectrometer:
DAV5 V3.01 Raman Spectrometer uses an Aries 532nm 150mW CW/DPSS Green Laser
Cost - $ 199.99 US
The DAV5 V3.01 3D Printable Raman Spectrometer
DAV5 V3.01 Raman Spectrometer - $ 657.35 US
... Read more »
Spectral-Database-1.0-master (3).zip
This is the latest installment to my spectral database on github, there are now 125 absorption and emission scans. PNG pics are also included relative to the particular studies being done.
x-zip-compressed - 3.19 MB - 04/15/2017 at 14:16
Download
Spectral-Database-1.0-master (2).zip
4 new raw data files added, parts 1 and 2 from; Effects of Solvents on Fluorescence of Biological Dyes done on 8/26/2016
x-zip-compressed - 1.70 MB - 04/13/2017 at 15:00
Download
Spectral-Database-1.0-master.zip
This is the begining of my spectral database collection project over at Github
x-zip-compressed - 1.49 MB - 04/13/2017 at 08:31
Download
FIXED cuvette holder FEB 2 (Meshed).stl
*UPDATED: 2/5/2017 re-designed the Raman filter slot again, I found a less expensive Raman edge filter at Thorlabs that will work just fine, so the new slot is now 25.5 x 4.45mm and 39mm deep so the filter is at 12.5mm center line.
Standard Tesselated Geometry - 1.41 MB - 02/05/2017 at 13:12
Download
25 x25 x 6mm diffraction hologrphc done (Meshed).stl
UPDATE 1/17/2017:Holographic Diffraction grating mount holder. Dimensions; 27 x 27x with a 6mm width slot for grating mount. Inlayed 2mm at slot bottom.
Standard Tesselated Geometry - 597.64 kB - 01/17/2017 at 23:07
Download
View all 12 files
  • 1 × Cuvette Holder W/ Exit slit plate Dimensions; 45 X 45 X 25 mm / 3D-printed by Sculpteo
  • 1 × Round standoff M3 Nylon 13mm
  • 1 × Round standoff M3 Brass 11mm Used for enclosure
  • 1 × Machine screw Pan head phillips M3
  • 1 × Machine screw Pan head phillips M4
View all 14 components
  • *Update* V3.02 Raman Spectrometer Solux Lamp spectrum Scan
    David H Haffner Sr • 11/30/2017 at 15:06 • 0 comments
    This is a critical alignment test before I install the new 1st collimation mirror and 1200ln/mm diffraction grating, I wanted to make sure that my mirror and diffraction grating placement are in their proper aligned positions.
    The results are very good, so now I can be assured that the CCD's angle will be correct as it relates to the 2nd mirror and I will get the full spectrum on the CCD.
    These are the main 3 points of reference for this particular broadband spectrum (Solux lamp 4700K, 50W, 12vdc.) The Original spectrum from the Solux factory has the points of reference at 529, 701 and 1013nm.
    MIne is a little off right now because of my 1st mirror ( a special silver coated reflective flat mirror [99% reflectivity]
    not bad, but not good enough. The 1st mirror will be like the 2nd mirror, 1'' convex focusing mirror with a focal length of 100mm, and a working distance of; 3.54''.
  • CCD Detector 8-Bit TCD1304 Finished
    David H Haffner Sr • 11/09/2017 at 19:24 • 0 comments
    The TCD1304DG CCD detector for the DAV5 3D printable Raman spectrometer is finished, this CCD was a collaborative effort between myself and Dave Allmon. His design, my work and experimentation and finally his debugging and final PCB product have made this beautiful detector possible :)
    All details and specs can be found here: Fast 8-Bit CCD detector TCD1304
    CCD detector MAIN SCH
    Motor Control Interface
  • Motor Control Targeting Interface Mega 2560 V3.02
    David H Haffner Sr • 11/06/2017 at 14:57 • 0 comments
    *UPDATED Turret Control Program:
    /*When you turn the rotary encoder clockwise, the angular displacement is increased;
      when you turn it counterclockwise, the displacement is decreased.  If you press the switch on the rotary encoder, the readings will return to zero  Original code can be found here://Email: support@sunfounder.com  //Website: www.sunfounder.com  //2015.5.7/
      **Although this is in the Public Domain, I have created a "Creative Commons License" for this work,  the reason is, I made enough mod's and put in many hours of work to make this program accurate and  efficient as part of a much larger project @ Hackaday.com, the DAV5 3D Printable Raman Spectrometer.
      There are individuals there that have engaged in "intellectual theft" and not bothered to give credit  where credit was certainly due, so I believe I have no choice but to go this route since I have put in  2 years of very hard work on this.**
      The creative commons license can be viewed here: http://thequantumhub.blog/
      I have modified this code to accomodate my diffraction grating turret for  the DAV5 Raman 3D printable Spectrometer (http://hackaday.io/project/18126-dav5-v301-raman-spectrometer)
      Written and modified by David H Haffner SR 10/18/2017
      Full project details can be found here @ https://hackaday.io/project/18126-dav5-v301-raman-spectrometer
      This program is also in the public domain @ https://playground.arduino.cc/Main/KY-040EncodeUsedToControlAHolographicDiffractionGratingForARamanSpectrometer
      It utilizes a KY-040 encoder to control the holographic diffraction grating  at steps of 1 & -1, each pin (SW,DT and CLK) has a 0.1uf ceramic cap for decoupling,  this is more effective for rise time spike removal. There are 2 LED indicators, a red for encoder step movements and  a yellow for home position indication that will light when the shaft docks back to a pre-determined home position,(I'm still  working on it,)This is accomplished by adjusting both "StepsToTake and RotoryPosition," they are in direct relation to steps per revolution  of the internal shaft according to the type of stepper motor you are using (mine is a 28-YBJ-48).Fine tuning can be done by playing between  STEPS (per revolution of shaft) and Speed, this will ensure smoother transitions when turning the encoder for fewer misses.  If you press the switch on the rotary encoder, the readings will return to approx zero 0.21 percent error.  a 0.1uf ceramic Cap is placed @ PinSW (4) to + (the reason for the pin pullup).
      The formula in this sketch is derived from the specifications of the 28-YBJ-48 stepper MTR  which has a step angle of 5.62deg, to convert degrees into nanometers you 1st have to convert  degrees into radians {RAD}, we do this for the 1st specification which is 5.625deg = 0.09817474 {RAD}  this # is then divided by the equitorial radius of the Earth which is 3,963.2 miles.
      Our new value would then be, 0.000024772 miles (NM), convert this into centimeters, (3.986667)
      Now with this #, we multiply it by the constant of 1 (NM) = 1609344e+012 miles *(Variable)/ 7  gives us our converting factor from encoder position count to nanometer counts.
      Sketch uses 9202 bytes (3%) of program storage space. Maximum is 253952 bytes.  Global variables use 519 bytes (6%) of dynamic memory, leaving 7673 bytes for local variables. Maximum is 8192 bytes.
    1. include <digitalWriteFast.h>
    2. include <math.h>
    3. include <Wire.h>
    4. include <LiquidCrystal_I2C.h>
    5. include "Stepper.h"
    6. define STEPS 64 // Number of steps for one revolution of Internal shaft
    // 2048 steps for one revolution of External shaft LiquidCrystal_I2C lcd(0x3F, 2, 1, 0, 4, 5, 6, 7, 3, POSITIVE); // Set the LCD I2C address
    double double_x = 3.986667 * 1.609344e+012 * 0 / 7;
    volatile boolean TurnDetected; // need volatile for Interrupts volatile boolean rotationdirection;...
  • Final CCD Circuit Board TCD1304DG Verified Working...Ready for final steps
    David H Haffner Sr • 10/13/2017 at 14:13 • 0 comments
    Final board design (needs 1 minor track change from C9 & C13 on PCB, could have been my fault upon manual component placement.) All in all, this baby works :)
    The plots below will show a red laser pointer (<5mW) aimed approx. center on the CCD, NO diffraction grating but I used a 2nd order regression formula table I made on Excel to calibrate the pixels into wavelength (nm).
    This is the un-calibrated plot, it has been corrected for Intensity sensitivity (using an official data set from the Solux corp. technical website/Solux broadband spectrum-4700K ( 260 - 1100nm)
    This is the summary output from Excel using 8 "true" wavelength peaks from an Hg calibrated lamp spectrum. The R squared value is not quite were it should be ( .999) but that's because there was no diffraction grating and I was holding the Laser pointer with my hand.
    Still very close, this is just an initial example, making sure the circuit board is functional. The laser pointer states 650nm, but I am getting a calibration of 700nm, probably is about 650nm when I get it in the enclosure with all the optics in place, it should register normally.
    This is the FWHM value data
    This is how I have the board designed so the USB cable enters the FTDI from the rear and power also enters from the rear and connects to the Arduino Mega 2560 mounted on the side of the enclosure.
    Finally, a front view of the PCB, everything is tightly placed together, I found out something very interesting since this is the 5th re-design, the ADC does NOT like to be too close the LM324 (opamp) I'm sure I could have solved this problem if I would have taken a little more time and perhaps followed the recommended PCB design on the data sheet for the LM324 but it also required a low pass buffer and I just was not in the mood for so much work, sorry.
    Anyway I am excited that it finally works well and can't wait to get it in the enclosure with all the fancy optics!
  • Consistency is the key to success
    David H Haffner Sr • 09/29/2017 at 13:48 • 0 comments
    As my project reaches its final phases, I am reminded of the basic tenets of the scientific method:
     The scientific method requires making observations, recording data, and analyzing data in a form that can be duplicated by other scientists. In addition, the scientific method uses inductive reasoning and deductive reasoning to try to produce useful and reliable models of nature and natural phenomena.
    One may not really pay that much attention to this when building projects as we do but the same applies as well, especially for a project such as this.
    This flow chart depicts the four basic steps in this process, sometimes I think we forget how to properly utilize this chart when rushing to complete a project to gain a little popularity and monetary gain in the market place and what happens is, the final product becomes somewhat sub-standard at best and perhaps a bit in question as to quality standards of results.
    A shinny object with many bells and whistles but no substance under the hood, a good project does not make! A project with a trail of evidence acquired through the scientific method is best.
    Case in point;
    The two resolution tests done today 09/28/2017, were carried out with two laser pointers, a 532nm green laser <5mW and a 405nm violet laser <5mW. The point of the experiment was to test my CCD driver (TCD1304DG,) detector’s resolution as it stands at this moment, and check for consistency. Even though I am not using any diffraction grating in my test platform I have a dichroitic mirror with a cutoff at 650nm that works like a beam splitter, quite adequate for testing purposes.
    The setup is simple, the entrance slit is 2 Gillette razor blades placed at 0.25mm width, distance to 1st mirror (silver coated @ 99% reflectivity,) is 70mm, the beam then travels @ 45 degrees to the dichroitic mirror and is reflected to the CCD @ 39 degrees (blaze angle).
    In this way I can capture 85 percent of the beams wavelength and tame the Ariel disk effect without having to use any additional filters or optics. A rather unconventional approach I know, but it works.
    This test platform allows me to utilize this circuit as my Beta board, meaning it’s not the final product. This lets me test its functionality in real time and lets me make real time adjustments and design changes as needed. The final board is already to go as soon as I do a few more test runs on this one and I’ll send off the files to get the PCB’s made.
    The new dimensions will allow for an excellent fit in the main spectrometer’s enclosure, I was able to reduce the circuit board to 3.5” X 3” (88.9 X 76.2 mm).
    Step number two in the scientific process here as it relates to spectroscopy is, spectral calibration. What I mean by this is, I can’t just drop a sample in the cuvette holder and shoot an image and call it a day, post a spectrum and claim its what I said it is, it doesn’t work that way.
    There are standards of practice when it comes to qualitive chemical analysis.
    • Use Traceable Standard Reference Materials
    Reference material uses include validating methods that help ensure accurate data from individual test runs, calibrating instruments and assessing analyst proficiency. In the United States, a NIST standard reference material is considered the “gold standard” for that material.
    2.    Run in Duplicate
    The purpose of duplicate (sometimes triplicate) testing is to add to the confidence that the test run has produced good data for the test object. Replicate data that agrees is a good measure of method reproducibility but does not prove data accuracy (validity). If the same test run includes a reference material, then the confidence in the validity of the data for the test object is significantly raised. If the object’s replicate test data is not in agreement, one or more of the data points may be invalid; the object should be retested and/or the procedure should be reviewed....
  • 1nm Resolution Test Successful 8-Bit CCD Detector TCD1304DG
    David H Haffner Sr • 09/26/2017 at 15:17 • 0 comments
    Well, this may be the ugliest test platform I have ever made but it sure worked, I had to modify my circuit board @ R8 going to pin#3 at the opamp [LM324], by running a bypass from the C9 [200pf cap,] to pin#3 @ the LM324, some strange ground problem I can't quite figure out yet from manually placing the components on the PCB design circuit in DesignSpark.
    Anyway the bypass works like a champ, I have a very low noise floor [SNR] and a very good signal :) I used a RED 650nm laser pointer as my test spectrum to determine resolution, used NO diffraction grating, only a 0.25mm razor blade slit and my silver coated mirror.
    Straight to the point, the data is UN-calibrated, for now, but initial results are beautiful.
    Data first...
    Figure 1 is the 1st spectral image starting from 1500 - 1726 pixels (X axis,)
     Figure 2 shows the spectrum cutoff @ 1700 - 1726 pixels (X axis,)
    Next figure (3) is the FWHM data set,
    Here we can see peak position @ 1717 pixels (nm) with A Full Width Half Maximum (FWHM) of 1.0547nm, most excellent, especially since I was only using a makeshift entrance slit made from two gillette razor blades @ a width of 0.25mm.
    Below is my test platform I made just for this purpose:
    Next image is a closer look at the circuit board:
    And finally a better look at the angle of the mirror in relation to the detector:
  • PLX-DAQ v2.11 Dual Trace Transfer to Excel
    David H Haffner Sr • 09/20/2017 at 15:28 • 0 comments
    This is my second installment on using PLX-DAQ as an easy method of sending data from an Arduino or any AVR related MCU to Excel. Here I will be showing you how to process the data into a graphical chart for plotting it’s values…the easy way 🙂
    Download PLX-DAQ v2.11 here: http://forum.arduino.cc/index.php?top...
    Code can be found here: http://playground.arduino.cc/PLXDAQ64...
  • Updated PCB for 8-BIT CCD Detector (TCD1304DG)
    David H Haffner Sr • 08/25/2017 at 09:28 • 0 comments
    Updated 08/25/2017
  • Adjustable Mechanical Slit 3D-Printed for Raman Spectrometer
    David H Haffner Sr • 08/14/2017 at 16:05 • 1 comment
    This was a design I worked on back in December of 2016, I took the inspiration from a ThorLabs design, which gave me the idea to make one using a mechanical micrometer a means to manually adjust the entrance slit in increments of my choosing.
    Namely, in the range of 0.005/mm, it works on FreeCad v0.16 using the animation simulation, but I am still trying to work out some tolerance issues that are plaguing this design, mainly because of the 3d printing material and techniques.
    I'm going to upload the mesh/.stl files anyway in case someone can tweak it and make it better, don't worry there is no license or anything, as all my work here is provided freely, just relying on an honor system of giving credit where credit is due.
    The blue hexagon, are 2 Gillette utility knife razor blades, which I examined the edges with my microscope (100x) and the edges are surprisingly even, with fewer jagged edges than normal blades, so both edges come together with a very tight fit.
    I have the mechanical drawings available if anyone is interested, just let me know and I will email them to you.
  • *Data Update* 8-Bit CCD Driver Circuit (TCD1304DG)
    David H Haffner Sr • 08/03/2017 at 19:50 • 0 comments
    Well I've worked out some noise issues concerning the 8 bit ccd design, I'm glad I got to play around with it on the breadboard before ordering the PCB. I was able to make some needed tweaks in DesignSpark and update the gerber files and now they are getting made @ PCBWay.com :)
    I am confident in it's performance, even @ 8 bits, I did learn a few things though about working with CCD's, one is, thermal noise, yeah, even with NO power its still there, and on the breadboard it has little room for any air circulation underneath, so I had to compensate for this "dark current" by taking at least 5 dark frames and comparing the frame with the least amount pixel depth and then averaging the other frames together at their higher RMS #'s and subtracting that from the low pixel numbered frame and that gave me a pretty good representation of the most stable dark current values.
    I did the same procedure with the sample frame, I used my Solux 50 Watt Halogen lamp (4700K), which uses a 12vdc regulated power supply @ a distance of 52 cm and a slit width of 0.5 mm and slit length of 16 mm.  The biggest reason for following the same procedure with the sample frames is because of what is called "shot noise,"
    Which is caused by the random arrival of photons.  This is a fundamental trait of light.  Since each photon is an independent event, the arrival of any given photon cannot be precisely predicted; instead the probability of its arrival in a given time period is governed by a Poisson distribution.  With a large enough sample, a graph plotting the arrival of photons will plot the familiar bell curve.
    Shot noise is most apparent when collecting a relatively small number of photons.  It can be reduced by collecting more photons, either with a longer exposure or by combining multiple frames.
    I chose the multiple frames option because I am not capturing any live spectra, as one would if you were imaging a galaxy for example, so I used an integration time of 0.50 ms long enough to get the shot but short enough to not introduce unnecessary noise caused by longer exposure times.
    The first plot we will look at is figure 3, because its the raw spectrum from 260 pixels to about 1200 pixels in length. I processed the CCD data using Spectragryph 1.2.0 designed and written by Dr. Freidrich Menges PhD. and is now certified by the NIST. I already subtracted it from the dark frame and flipped it right side up using this LOG formula in excel; =(10*(-B2))*100. This will convert the Y values to -LOG so the spectrum will revert to a transmission spectrum, because negative numbers do not make sense in this case.
    The next plot figure 2, is the spectrum cutoff at 385 - 800 pixel counts. This was the reason for the 16 mm slit length, the only way to manually control the wavelength range was to simulate an actual spectral image on the CCD substrate without a diffraction grating.
    In figure 2 you also notice, how I calibrated my Solux spectral image to correspond to a master reference spectrum, a Solux spectral plot I downloaded from Solux Technical Data and Graphs
    Here is the spectral resolution data for this spectral plot below
    Keep in mind that, @ 1 and 2 nm it's not the best but we are working from a breadboard after all, with no diffraction grating or optics!
    Finally I put it all together in figure 1, best for last, and introduced a feature in Spectragryph 1.2.0 called "Spectropic view," where I can align the plot to fit a particular wavelength range and image the spectral lines.
    The next steps are clear when the PCB boards come in and then the final phase of testing can begin :)
View all 119 project logs
  • 1
    Step 1
    These are the build instructions for my LS-532 CB/fiber optic laser collimation tube;
    New Holographic diffraction grating holder assembly.W/drawings:
    The newly re-designed parts came in so I began to prep them and wanted to include them here under build instructions;
    These are the first preliminary drawings and CAD blueprints depicting several aspects in the assembly design and by no means is this the final assembly steps;
  • 2
    Step 2
  • 3
    Step 3
    10X Lens Holder Base assembly and holder design W/ slide rail feature:
View all 13 instructions
G.Vignesh wrote 08/24/2017 at 15:34 • point
This is great! That too with 80% 3d printed objects!
Cool stuff sir!
David H Haffner Sr wrote 08/24/2017 at 15:47 • point
Thanks G.Vignesh, the 3D parts actually are the most difficult to design :)
G.Vignesh wrote 08/24/2017 at 15:58 • point
One question sir! What platform did you use for 3d modelling? Fusion 360 or?
David H Haffner Sr wrote 08/24/2017 at 16:17 • point
I use FreeCad v0.16 and POV-Ray v3.7 for Windows to do the rendering part.
G.Vignesh wrote 08/24/2017 at 18:50 • point
Great!
Fredgbh wrote 07/15/2017 at 08:14 • point
Please don't send the uC board for fabrication, it needs some work, some aspects are sub-optimal (that's code for something impolite).
I think you need to start again:
Place your analog section all together, away from the digital bit. Short traces, consider crosstalk, and having a separate analog Gand
Think of the xtal as analog as well, short traces, away from signals of all species.
Do the placement, then route the power, separate filter for analog power
Only then do the signals, if the power and clock aren't right the rest will be a belt made of watches.
Apologies for the slight snark, you are clearly intelligent and have got off your a$$ and done Something, kudos!
I would like to help avoid the frustration when signal integrity doesn't happen
David H Haffner Sr wrote 07/15/2017 at 12:52 • point
Hell Fredgbh, no offence taken...yet, which circuit are U talking about? The 16 bit of 8 bit? You have to let me know so I am clear what we are talking about first, Thanks :)
Fredgbh wrote 07/15/2017 at 17:05 • point
Schematic says 8 bit.
I have no sense of what analog performance you need to reach and can't correlate the quality of your layout with the likely performance.  What I can see is that aspects of the layout are sub-optimal. First the schematic errors:
- Both sides of C2 are connected together to GND - easy to see on the layout, not so easy on the schematic
- Common point of C4 and C5 needs connecting to GND
- C3 is connected backwards, the schematic symbol is pretty unclear but again obvious in the layout with the + terminal connected to GND
- LM324 unused amplifier inputs left floating
- Not an error but the CCD datasheet appears to suggest connecting the NC pins to GND - always good practise not to have floating bits of metal
Now the layout - if it works well enough for your application then no problem but there's room to improve it.
Placement:
- Minimise the length of the high impedance analog traces, get the components connected to the analog inputs of the LM324 near to the chip. Currently these sensitive nodes are dragged across the board under the uC
- Place components to minimise net lengths, eg D1 near the power input, anything else relatd to the regulator near to it.
- I'd recommend using 0.05" for GND and Power grid, its not just GND that needs a low impedance trace
- Separate analog gnd and power connected together at one point so that digital currents don't flow in the analog GND
Cosmetic - none of these will stop anything working - feel free to ignore
- The schematic isn't just a means to an end, its the documentation for the circuit, current one is hard to read even with so few components - review some others, perhaps more complex ones to get an idea.  This won't stop it working but makes it harder to debug.
- There are refdes in all four orientations - harder to read
- 45 degree trace angles are not necessary but give a more pleasing and compact result
Muriel Green wrote 07/04/2017 at 21:12 • point
Reading the news today and they mentioned raman spectroscopy! http://www.bbc.com/news/science-environment-40494248
David H Haffner Sr wrote 07/05/2017 at 07:11 • point
What a fascinating article Muriel, thank you :)
David H Haffner Sr wrote 03/09/2017 at 12:26 • point
@fl@C@ , I did not mean to offend your project, my thoughts about your methodology are just my opinion, not a rant about my superior intellect. You made a few bold claims about your devices capabilities and did not provide enough documentation to support them, that's all I was getting at.
In the scientific spectrum, there is always a clear trail of evidence that others should be able to follow in order to verify claims made, and to all experiments.
Your project is quite extraordinary,
no doubt, but there are several flaws in some of the capabilites that you claim that it can do. As far as utilizing some of your design concepts, very few, the 3D prints are my own design, the Czerny-Turner configuration is just good science, the CCD detector and MCU, are NOT your design, so get over that.
The .src code I am using is not yours either. When I talk about "full" documentation, that's exactly what I mean, not only the mechanical side of things, but the chemistry and spectroscopy side of things also...The process, that's what is important. Yes, the long boring techy stuff, the stuff like, what standard did you use to calibrate and verify your spectrometer? Did you use holmium Oxide?
Because when all is said and done, that is what I'll be using (and I'm going to have to make the $$ sacrifice of about $450.00US,) to verify that my spectrometer is on point and ready to go. I'm not recommending that anyone else has to do this, I'm doing it because I want this project to reflect my commitment to excellence.
Let me explain once more this projects details:
The DAV5 V3 Spectrometer will be the only project build here on Hackaday in its category, with ongoing, full performance and specifications documentation, this project will also be at least 95% 3D printed! The main goal for this project has 4 points:
1) To build a rugged and reliable low cost spectrometer with 3D printable parts.
2) A spectrometer capable of producing consistent and professional quality results in the field, home or Lab
3) A device with upgradeable capabilities
4) Every part of this project will have it's .STL file (for 3D-printing,) available for download.
**This project will also have full professional/analytical and chemical documentation.
**Also as an aside, I have owned my Aries 532nm Green laser (150mW) for over 4 years,) so I did not rip that idea off either.
Loki wrote 03/03/2017 at 12:07 • point
Especially with many cell phone cameras enabling access to RAW format images from their cameras (iPhone 6 and newer, many androids), is there any reason not to skip the CMOS all together, and use a cellphone camera; and also use it for the image processing?  Essentially, just have the light source, slit, diffraction, and raman filter in this device, and design it to mate to a cell phone camera?  That would give 8MP+ of resolution.
David H Haffner Sr wrote 03/03/2017 at 12:23 • point
Well Loki, the problem with that idea is, cmos has a spectral limit of 1100nm, (metal oxide,) Raman spectroscopy requires a set up like the Czerny-Turner configuration. Also you need a stable laser source coupled with an appropriate bandpass filter.
Loki wrote 03/03/2017 at 12:54 • point
I think that's what I just said.  When I said "device", I meant external device: all the parts of this project (or the Montoya one), except for the CMOS camera and image pipeline.
David H Haffner Sr wrote 03/03/2017 at 13:01 • point
"and design it to mate to a cell phone camera?  That would give 8MP+ of resolution."
you stated the above though, and that's what I was going by, if I misunderstood sorry.
Loki wrote 03/03/2017 at 13:07 • point
No, that is correct.  Instead of using an older DSLR such as is used in the Montoya design, use a modern cell camera.
Could you explain more about why the CMOS spectral limit is a problem?  perhaps that's what I don't understand, in light of a CMOS sensor being successfully used in the Montoya design.
David H Haffner Sr wrote 03/03/2017 at 14:18 • point
It all comes down to photons interacting with silicon. cmos if you are not familar, are 1000's of microscopic photo diodes, photons with an energy above 1.1eV will not necessarily interact with silicon. Their probability interaction depends on how much higher than the 1.1eV band gap the energy of the photon is, this defines an absorption coefficient  [cm-1] depending on the wavelength and the type of material.
This absorption coefficient defines which percentage of the photons entering a one centimeter material will be absorbed.
To make it even simpler, silicon as a material substance, will only by its own nature, absorb so many photons no matter what, and it just so happens that the limit is approximately 1125nm.
So, as long as you are willing to stay around the 400 - 700nm range in the spectrum you'll be fine...see my point?
Loki wrote 03/03/2017 at 14:32 • point
If we're going for stokes raman:
"Raman signals excited by a 532nm laser are distributed in the visible range, where the response is best for most silicon-based CCD chips. Meanwhile, Raman signals from 785nm systems fall within the NIR range (750-1050nm), where the response is still relatively good. For 1064nm, however, typically there is no response from the CCD above 1100nm" (http://bwtek.com/wp-content/uploads/2015/07/raman-laser-selection-application-note.pdf).
As long as we use a 532nm laser, the resulting stokes radiation we need to detect stays in the visible range - which would seem to allow for the use of a CMOS camera.  Also, to get around issues with low signal intensity, why are people sticking with 150mw range lasers, when so much higher power lasers are available for approximately the same cost?
David H Haffner Sr wrote 03/03/2017 at 15:07 • point
Yes, you are correct about that, as I stated; as long as you stay within the boundaries of 400-700nm, a 532nm CW laser at at least 150mW will work. The reason I changed up my design scheme was that at 532nm for that particular set up would not have been very effective for analytical Raman spectral work of liquid mediums.
This type of Raman spectroscopy works best at the reflectance level, and I just didn't want to drastically re-design the whole system to accommodate it. Too many more parts and 3D printed parts, more headaches :(
Ok, typical 785nm Raman spectroscopy is just NOT feasible for a cmos camera, after 700nm the spectrum gets very "scraggly," unpredictable and very unusable. http://www.aphesa.com/documents.php?class=wp - click on the pdf file that says; "Spectral Response of Silicon Sensors"
finstp wrote 02/28/2017 at 21:57 • point
There is a very simple design for a home-made 3D printed Raman spectrometer in the literature, complete with very nice example spectra.
See 'A Homemade Cost Effective Raman Spectrometer with High Performance' by EH Montoya, A Arbildo and OR Baltuano published in Journal of Laboratory Chemical Education 3(4) p 67-75, (2015) doi:10.5923/j.lce.20150304.02. This could easily be coupled with the Spekwin32 software (see here on Hackaday), which can extract a linescan spectrum from the digital camera image and you have all the benefits of professional software too!
Crypto [Neo] wrote 01/18/2017 at 01:32 • point
This is actually really really awesome. Battelle is a defense contractor here in the states, they sell a biohazard detector that is pretty large for a lot of money, this thing can be made for a fraction of that cost and easily applied to do biohazard detection. Great work!
David Challener wrote 01/13/2017 at 19:19 • point
I have been looking at Raman designs, and they all seem to have beam splitters in them. but I don't see one in your design. I have never been really sure why there was a beam splitter in the first place, so perhaps it isn't needed, but I don't understand why anyone would reduce the amplitude of the reflected light by a factor of 2 if it were not needed.  Additionally, Is more lines better or worse? (I can get a good 2500 lines per mm for $26).
David H Haffner Sr wrote 01/13/2017 at 19:56 • point
Hey David, first, the more lines per mm the higher the spectral resolution, but will "reduce" your wavelength range, so when choosing a higher order grating you have to make sure you know what wavelength your working in.
1200 lines is perfect when using Holographic gratings or high quality ruled blaze angled gratings, the DVD piece is fine at 1540 lines, mostly because some of those lines are not placed so perfect during manufacturing, but there are enough good lines that are parallel where the diffractive properties are just fine, (hence the Schott glass color filter.)
Second, beam splitters, especially the cube type, are used to separate wavelength and power distribution of the laser's collimated light, in the case of Raman spectroscopy, it is used because in most set ups, there is an adjustable diffraction grating that can be rotated to select a particular wavelength.
I don't have to use it, although it certainly would improve the overall sensitivity of my system but then it will start to become very complex and costly and since I am only concerned with the region between 500-4000 cm -1 (546.5 - 675.6nm,) I am confident all will be well.
I know it may seem a bit strange why you would want to attenuate the incoming light beam, but depending on the application, you may want to only have 45% transmission 55% reflectance or aligned for maximum p-polarized transmission (99% reflected - 1% transmitted.)
David H Haffner Sr wrote 01/13/2017 at 20:05 • point
Also I wanted to add a quick note, mine has a secret weapon; 532nm, 12.5mm Diameter, Raman Edge Filter. http://www.edmundoptics.com/optics/optical-filters/longpass-edge-filters/raman-longpass-edge-filters/86225/
Ryan White wrote 01/23/2017 at 11:07 • point
There ya go, that's what's up, but you've gotta hold your laser on that wavelength, and your laser needs to be single mode, what's your source?
David H Haffner Sr wrote 01/13/2017 at 08:38 • point
Hey Ted, for an application such as this one, I would use a UV Reflective Holographic Grating, 1200/mm. Cost is around $131.00US, depending on the source and quality.
Also I wanted to add that, Holographic gratings have a low occurrence of periodic errors, which results in limited ghosting, unlike ruled gratings. The low stray light of these gratings makes them ideal for applications where the signal-to-noise ratio is critical, such as Raman Spectroscopy.
I am hoping that the Schott glass color filter placement with the diffraction grading placed before the spectral image strikes the detector, will eliminate the stray light and most ghosting effects, if not, I will be forced to adapt and use the holographic grading, which will cause a slight re-design and significant up-cost.
David Challener wrote 01/12/2017 at 22:56 • point
So next question - why are you using a DVD for your diffraction grating?  I see them used in very low cost spectrometers, but they clearly aren't quite parallel lines.  Does this matter?
David H Haffner Sr wrote 01/12/2017 at 23:53 • point
Hey David, good question, I'm using the 4.7G DVD piece with 1540 lines per mm. Spectral resolution is more important than how straight the lines may be on the diffraction grating, I've been researching this aspect for over a year now and I know that it does work, you just have to be willing to be patient and rotate the grating carefully until the spectral lines are straight.
I explained how to do this in a previous log and in some research notes on Public Lab. That is also why I cut the piece 18 x 18mm square and placed each corner on the flat top end of the Schott glass filter locking sleeve, so it can be rotated as you are watching the spectral lines in live capture mode on screen.
I use a 2300K 13W CFL (compact fluorescent light) to calibrate the diffraction grating piece because of the mercury lines.
I hope this explanation helped!
David Challener wrote 01/11/2017 at 17:14 • point
I am interested in building one of these.   Is it ready to go?
David H Haffner Sr wrote 01/11/2017 at 17:29 • point
Hey David, everything is ready except the Raman longpass edge filter, I will not be able to purchase it until sometime next month, I included it on the bill of materials because it certainly needs to be there.
The rest of the device is performing beautiful in the UV/VIS range-300-800nm. It will do absorption/fluorescence and UV as it stands at this moment.
PointyOintment wrote 01/10/2017 at 09:22 • point
Even if the category is as narrow as "open-source 3D-printed Raman spectrometers", there's already another project in it: #ramanPi - Raman Spectrometer
David H Haffner Sr wrote 01/10/2017 at 11:38 • point
Hey pointyointment, yes that is true, but the Ramanpi project's main documentation is mainly centered around the 3D printing of their design and on their concept of home automation. Proving their concept is very questionable for such bold claims.
They do not have sufficient scientific documentation validating the Raman aspects of their device. The process for doing so is very meticulous and precise and must be done with caution, scientific claims must be able to stand up to intense scientific scrutiny by your peer community and not be prefaced by popularity.
My project has intentional and methodical steps and protocols in order for clear validation, when I am ready to capture Raman signals that will be a whole separate set of steps and protocols.
Also, my project includes cost analysis for full transparency, in order to keep the projects main goal under $700.00US. The average cost even for a 3D printed version of high sensitivity and precision can cost upwards of $2500.00 to 3000.00US Total.
Ted Yapo wrote 01/13/2017 at 00:35 • point
Do you know approximately how much cost it would add to the BOM to use a commercial replica grating instead?
fl@C@ wrote 03/09/2017 at 11:00 • point
I would have hoped that you would have contributed to, instead of insulting a project that clearly inspired not only the name of this project, but the design as well..  Although, after reading your replies to most of the comments here.....It doesn't sound likely that your input would be very useful when taken as a member of a group, rather than you taking credit and imposing your superior knowledge on everyone.  If your understanding of ramanPi is limited to the thought that it is centered around home automation, maybe you should review the documentation again.  It's quite interesting that you seem to have chosen many of the same components, for such a questionable concept...
Jarral Ryter wrote 01/05/2017 at 21:28 • point
yes raman has some sort of light scattering by shining laser onto the sample.... maybe you made a fluorescence spectrometer. Which is still cool.
David H Haffner Sr wrote 11/24/2016 at 14:47 • point
Hey Ryan, that's a good question, and I'm glad you asked it. Most Raman spectrometers are usually quite large and have very complicated optical designs, but the basic operating principle is still the same, resolving power of the optical mirrors and lines per mm on the diffraction grating.
A raman spectrometer can be composed of either a CCD or a CMOS detector, I have worked for a year researching the cmos type, and that is why I am using it, (cmos,) I am experimenting with the use of only one mirror, the square silver coated focusing one. The optical resolution of the spectrometer is determined by the input slit size and the optics inside the spectrometer. The best possible resolution that can be obtained is the diffraction limit, which is the resolution obtained with an infinitely small input slit. The diffraction limit is related to the size of the beam width inside the spectrometer.
The wider the beam, the more grating lines the beam illuminates and, therefore, the better the resolution. This is also referred to as the resolving power of the grating.
I will be posting a more detailed explanation of the theory and concept of my design, and the mathematics to back it up, plus I have research to back up this design also. So Raman merely means the resolving power of the spectrometer to clearly define the spectral lines with precise definition for accurate molecular finger printing.
Ryan White wrote 01/23/2017 at 11:03 • point
OK, but, Raman spectrometers use high intensity laser sources and look for the signature in the Raman shift of the scattered light, right? So basically you're looking for very weak signals very close to, but not at, the wavelength of the light source.
Raman spectrometers use laser sources of very high spectral purity (i.e. narrow spectral line width, low phase noise, whatever you want to call it) and very narrow band filters, and as far as I understand it both are tunable, you don't seem to have either. "modern instrumentation almost universally employs notch or edge filters for laser rejection and spectrographs either axial transmissive (AT), Czerny–Turner (CT) monochromator, or FT (Fourier transform spectroscopy based), and CCD detectors." - https://en.wikipedia.org/wiki/Raman_spectroscopy
This, as far as I can tell, is not a Raman spectrometer.
Ryan White wrote 01/23/2017 at 11:05 • point
I've just seen your more recent comments, I'd love to see this working and I'd love to see how you're controlling the wavelength of the laser source.
David H Haffner Sr wrote 01/23/2017 at 11:43 • point
Ok Ryan, my laser is a CW (continuous wave,) 150mW DPSS, I have posted the specifications for it here many times, I have a 532nm CWL, 10nm FWHM, 25mm bandpass filter employed in my fiber optic laser collimation tube assembly, I have already stated in previous postings that I am waiting on my 532nm, 12.5mm Diameter, Raman Edge Filter, it costs $450.00US, I am not made of $$.
I never stated it was a Raman spectrometer YET, it will be by the time I am finished, I strive for perfection, not speed or popularity. I do NOT need a CCD to detect Raman signals.
Finally, I am not going to explain myself over and over again, when I have been quite clear in my research and documentation, not only about Raman spectroscopy, but spectroscopy and chemistry itself. Read my work.
Ryan White wrote 01/23/2017 at 11:48 • point
Good luck with it! I'm watching with interest because I know how hard this stuff is, not because I think you won't do it. The internet is bad at capturing my tone, and (as I'm sure you know) there's been a lot of bullshit merchantry around handheld raman spectrometers for all kinds of applications, hence my skepticism.
Ryan White wrote 01/23/2017 at 11:52 • point
Nice write-ups as well, looking forward to seeing the results (and building one!)
Ryan White wrote 11/24/2016 at 13:37 • point
How is this Raman?