Close
0%
0%

Pwrblock 323

Reliable power for DUT testing: USB-C input, Python API, open hardware. Compact tool for labs, factories, and developers’ desks.

Similar projects worth following
How can you ensure thousands of devices work flawlessly, both in the lab and in production? By testing at every stage. But powering DUTs (Devices Under Test) has always been a compromise. Cheap power bricks are often unpredictable and unsafe. Lab monsters are expensive, oversized, and 90% of the features are useless for automated testing. We got tired of this trade-off. So we’re building our own: a reliable, compact, and obedient power supply, made for automated testing.

A compact switching power supply designed for test automation and HIL setups.

  • Powered from USB-C PD (100 W)

  • Controlled via USB with a clean Python library

  • Built with Open Source Hardware & Software principles

Our goal is to create a new category of instruments: affordable, open, and designed specifically for testing.

Features

  • Output: 0–32 V / up to 3 A (96 W max)

  • Ripple: < 30 mVpp, < 2 mA RMS

  • Protections: OVP, OCP, OPP, OTP

  • Measurement resolution: 10 mV / 1 mA

  • Control: USB-B (SCPI, custom firmware mods), PLD connector for extensions

  • Expandability: LEDs for status, button for manual toggle, optional module with screen/knobs

Technical Specs

Electrical Parameters

ParameterValue
1InputUSB Type-C Power Delivery 100W
2OutputNot galvanically isolated
3Number of channels1
4Max output power96 W
5Output voltage0–32 V
6Output current3 A max
9Voltage ripple< 30 mVpp
10Current ripple< 2 mA RMS
11Load regulation (10–90%)±0.1% + 20 mV / ±0.1% + 5 mA
14Load recovery time< 200 µs
15Line regulation (±10%)±0.1% + 20 mV / ±0.1% + 5 mA
18Voltage overshoot (turn-off)< 100 mV
20Setpoint resolution10 mV / 1 mA
23Setpoint accuracy±0.1% + 30 mV / ±0.1% + 5 mA

Measurements

  • Quantities: Voltage, Current, Power

  • Accuracy: ±0.1% + 30 mV / ±0.1% + 5 mA

  • Resolution: 10 mV / 1 mA

Protections

  • Overvoltage (OVP) — configurable

  • Overcurrent (OCP) — configurable, trip < 10 ms at 2× load

  • Overpower (OPP) — configurable

  • Overtemperature (OTP)

Structure

  • USB Host - control device (PC, single-board computer, laptop, etc.),
  • DUT - load, device under test
  • Type C Adapter - any USB-C power source, e.g., mains, power bank, laptop, phone.
  • The MCU controls all parts of the unit. This makes it possible not only to control the source from a PC, but also to load presets according to a predefined algorithm and link the power supply to external events.
  • PLD connector - a universal connector that provides input and output power, USB control, microcontroller interfaces, and logic power.

Of course, this is a specialized power supply, and we do not intend to replace laboratory power supplies. Let's talk more about the application options.

Use Cases

In Test Stands

  • Automated cycling, no manual knobs or displays

  • Board-to-board or internal wiring power delivery

  • Multiple modules for multi-voltage DUTs

  • PC inside the fixture, USB control via PLD

On the Desk

  • Simple power setup for developers

  • Configured via console, Jupyter, or SmuView

  • One button, LED indicators, optional expansion with knobs & screen

Roadmap

  1. Collect feedback from the community
  2. Build and test prototypes
  3. Crowdfunding launch on CrowdSupply

...

 N. Series of compact open tools: USB HUB, multimeter, production programmer, signal MUX, logic analyzer, DAQ...

  • From Prototypes to Production: PwrBlock 323 is Live on Crowd Supply

    Ioan Larionov08/18/2026 at 12:29 0 comments

    A few months ago, we shared the first real hardware results from PwrBlock 323.

    Since then, we’ve continued testing the power stage, measurement and control functions, working on firmware and SCPI control, and refining the design for production.

    Today, we’ve reached the next milestone:

    PwrBlock 323 is now live on Crowd Supply.

    PwrBlock 323 started from a simple problem we kept running into while building automated test fixtures: powering the DUT is often much less automated than the rest of the test stand.

    A bench power supply works, but it sits outside the fixture and is easy to reconfigure accidentally. Generic DC-DC modules are compact, but usually lack a stable software interface, measurement, protection, and long-term availability.

    So we built PwrBlock to become part of the test system itself.

    The current design provides:

    • 1–32 V programmable output
    • Up to 3 A output current
    • USB-C PD power input
    • SCPI control
    • Python-friendly host tooling
    • Voltage and current measurement
    • OVP, OCP, OPP, and OTP protection
    • Banana outputs for bench work
    • Pin headers, UART, SPI, I²C, and digital I/O for fixture integration

    The idea is that a test script can configure DUT power, enable or disable the output, monitor voltage and current, and power-cycle the DUT without any manual interaction.

    That makes PwrBlock useful not only as a compact programmable supply, but as a building block inside production jigs, HIL systems, automated lab setups, and field-debug rigs.

    What happens next

    The core power, measurement, and control functions are already running on working prototypes.

    The Crowd Supply campaign is the next step toward finalizing the production design and manufacturing the first batch.

    We’re also building PwrBlock as an open platform. We plan to publish the hardware design files, firmware, mechanical files, and host-side software so it can be understood, integrated, repaired, and adapted instead of becoming another black box inside a test fixture.

    Every production unit will be calibrated and tested by Everypin in Finland before shipment.

    During the campaign, we’ll keep posting technical updates here: hardware tests, SCPI and Python examples, design decisions, manufacturing progress, and probably a few things that don’t work on the first try.

    If programmable DUT power is something you need in your own test setup, take a look at the campaign:

    https://www.crowdsupply.com/everypin/pwrblock-323

    And if you know someone building automated hardware test fixtures, please send it their way.

  • From Concept to First Hardware Test Results

    Ioan Larionov03/31/2026 at 13:35 0 comments

    PwrBlock

    PwrBlock is our open-source programmable power supply for test fixtures and electronics testing automation.

    It is no longer just a schematic or a render. We already have the first assembled samples, and they have gone through the first real tests. This article is a short update on what PwrBlock is, what we have already confirmed on hardware, and what we are improving next.

    Why We Are Building It

    The idea behind PwrBlock is simple.

    When building test fixtures, engineers often end up using either a full bench power supply with lots of unnecessary features, or a generic DC-DC module that works until it suddenly becomes the least predictable part of the setup.

    We wanted something else: a compact, programmable, open hardware block that can be embedded directly into a test stand and controlled as part of the system.

    PwrBlock is part of a larger plan to build a family of open hardware tools for electronics testing: power supplies, programmers, switch matrices, DAQ modules, and other building blocks that can be combined into complete fixtures.

    What Is Inside

    Article content
    Main components

    The current revision is built around a few key components.

    Protection: TPS25983

    • 2.7 V to 26 V
    • 2.7 mΩ integrated 20 A FET

    MCU: STM32G0C1CE

    • Cortex-M0+ CPU, up to 64 MHz
    • 512 KB flash
    • USB Type-C Power Delivery controller

    The main reason we chose this MCU family is the built-in Power Delivery controller. It lets us avoid using a dedicated USB PD IC.

    Measurement: INA229

    • up to 85 V
    • 20-bit, ultra-precise
    • SPI interface with alert output
    • integrated temperature sensor, ±1 °C

    This chip gives us accurate measurements of current, voltage, power, and energy at the output of the supply, without extra calibration. It also gives us temperature data.

    Power stage: LM51772

    • Buck-boost
    • I2C configuration interface

    This is the heart of the block. It can step voltage down, step it up, and be configured over I2C.

    First Confirmed Power Results

    One of the first things we wanted to validate was long-term operation near maximum output power.

    So we ran a simple test with these conditions:

    • Output: 32 V / 3 A
    • Input: laboratory power supply
    • Load: electronic load
    • Board condition: no enclosure, no heatsink
    • Duration: 3 hours under load

    The results were:

    • Input power: 103.64 W
    • Output power: 96.17 W
    • Efficiency: 92.7 > %
    • Heat losses: 7.47 W

    That was an important milestone for us. It means the project has already moved from concept to confirmed real-world operation close to the target power level. Although it's not correct to measure efficiency after the wiring.

    Article content
    Testing maximum power for 3 hours

    Thermal Behavior

    After three hours at around 96 W output power, we checked the thermal picture of the board.

    The hottest point on the top side was the inductor body at 80 °C. The DC-DC converter itself stayed at 71–72 °C. On the bottom side, the hottest area was the thermal path from the inductor pads into the inner PCB layers, reaching 86 °C.

    The inductor itself is rated for temperatures up to 165 °C, so we do not expect thermal problems from these results. Still, thermal chamber testing is definitely on the roadmap.

    Article content

    The main takeaway is clear: the block can already operate without overheating even without an enclosure. The enclosure test is still ahead, but it is already obvious where the next improvement should go. We will likely use a larger inductor, and thermal coupling into the enclosure through a thermal pad under the inductor area should work well.

    Article content

    Current Limiting at Low Current

    Voltage regulation over I2C works exactly as expected.

    We get:

    • 10 mV steps in the 1–24 V range
    • 20 mV steps in the 3.3–32 V range

    We also added a useful extra feature: analog fine adjustment. It works well for small offsets around the digitally programmed voltage and allows us to reach about 1 mV adjustment steps.

    This behavior is not explicitly described as a feature in the LM51772 documentation, but it works. We apply an offset to the ATRK input...

    Read more »

  • Crowdsupply campaign has started

    Ioan Larionov02/26/2026 at 20:41 0 comments

    Hardware projects die quietly: not because the idea is bad, but because uncertainty + scaling pressure hits too early.

    Crowd Supply gives hardware founders a rare thing — a path to first users and real shipping.

    Today we’re starting Everypin’s journey with our first open-source production testing building block, PwrBlock 323.

    If this resonates, please subscribe/support — we’ll deliver.

    👉 https://www.crowdsupply.com/everypin/

  • It's alive!

    Ioan Larionov12/23/2025 at 18:05 0 comments

    Finally, the very first Powerblock is assembled and works! What a miracle!

  • Almost ready for prototyping

    Ioan Larionov11/18/2025 at 11:52 0 comments

    Almost ready for prototyping. Fixing the last issues and choosing the right manufacturer.

  • block design layout

    Ioan Larionov10/23/2025 at 08:42 0 comments

  • Planning SCPI USB interface

    Ioan Larionov10/10/2025 at 05:51 0 comments

    We decided to implement the SCPI protocol via USB to control our Powerblock. It's time to decide which specific commands should be implemented. 

    Instrument Classification:

    • bf_keyword = DCPSUPPLY

    Base Functionality:

    • Capability request:

      • SYSTem:CAPability? - This query returns an <instrument_specifier>, example reply from power supply with additional functions: (DCPSUPPLY WITH (MEASURE&MULTIPLE&TRIGGER)). More about <instrument_specifier> at 1.4.1 SCPI 1999.
    • Output state control:

      • OUTPut[:STATe] {OFF | ON | 0 | 1} - change output state of device.
      • OUTPut[:STATe]? - queries the output state of device.
      • As stated in Command Reference, *RST sets the output state to off.
    • Source control:

      • [SOURce]:CURRent[:LEVel][:IMMediate][:AMPLitude] {<Current>} - When a power supply is operating as a voltage source, this command sets the current limit. When a power supply is operating as a current source, this command sets the output current.
      • [SOURce]:CURRent[:LEVel][:IMMediate][:AMPLitude]? [MIN | MAX] - query for setting of current (target, min or max limits).
      • [SOURce]:VOLTage[:LEVel][:IMMediate][:AMPLitude] {<Voltage>} - When a power supply is operating as a voltage source, this command sets the output voltage. When a power supply is operating as a current source, this command sets the voltage limit.
      • [SOURce]:VOLTage[:LEVel][:IMMediate][:AMPLitude]? [MIN | MAX] - query for setting of voltage (target, min or max limits).
      • At *RST a power supply is required to set the voltage and current to “safe” conditions. This is generally achieved by setting them to their values closest to zero.
    • Status Reporting:

      The device status system consists of the OPERation (information about normal device operation, current operations) and QUEStionable (signal quality information) subsystems. Each subsystem consists of 3 parts (Event, Condition and Enable register), ref. by NGE100 User Manual:
      • CONDition - this register is directly written into by the hardware or the sum bit of the next lower register. Its contents reflects the current instrument status. Read only register. Its contents is not affected by reading
      • EVENt - this register indicates whether an event has occurred since the last reading, it is the "memory" of the CONDition register. It is permanently updated by the instrument. Read only register. Reading the register clears it.
      • ENABle - this register determines whether the associated EVENt bit contributes to the sum bit (status mask). Read and write access. Its contents is not affected by reading.
    • STATus:OPERation[:EVENt]? - reading device condition via event (memory) register. More about bit field of this register at 9.3 SCPI 1999.

    • STATus:OPERation:CONDition? - reading device condition directly. More about bit field of this register at 9.3 SCPI 1999.

    • STATus:OPERation:ENABle? - reading status mask for operation register.

    • STATus:OPERation:ENABle {0-65535} - enables the bits in the status mask for operation register. More about bit field of this register at 9.3 SCPI 1999.

    • STATus:QUEStionable[:EVENt]? - reading signal quality register via event (memory). More about bit field of this register at 9.4 SCPI 1999. For a power supply, the bits of interest in the QUEStionable status structure are VOLTage and CURRent. When a power supply is operating as a voltage source, bit 1 (CURRent) shall be set. When a power supply is operating as a current source, bit 0 (VOLTage) shall be set. When the output is unregulated, both bits shall be set (for example, while the output is changing to a new programmed value).

    • STATus:QUEStionable:CONDition? - reading signal quality register directly. More about bit field of this register at 9.4 SCPI 1999.

    • STATus:QUEStionable:ENABle? - reading status mask for questionable register.

    • STATus:QUEStionable:ENABle {0-65535} - enables the bits in the status mask for questionable register. More about bit field of this register at 9.4 SCPI 1999.

    Additional Functionality:

    • Measurement capability:

      • For more details on...
    Read more »

  • Submitted to NGI0 Commons Fund

    Ioan Larionov10/06/2025 at 19:47 0 comments

    We’ve just submitted Powerblock to the NGI0 Commons Fund

    — a European initiative supporting open-source and open-hardware projects that strengthen the digital commons.

    Our proposal focuses on bringing Powerblock to a fully documented, manufacturable open-hardware design and preparing its public launch on Crowd Supply.

    The NGI0 program, coordinated by NLnet Foundation, funds small to medium R&D projects that advance open, trustworthy, and sustainable technologies. If selected, the grant would help us accelerate development, testing, and community documentation.

    In any case, we’ll continue to build in the open — all progress logs and design updates will remain visible here and on our GitHub.


    Fingers crossed 🤞 and back to building!

  • ILIM emulation test

    Ioan Larionov10/06/2025 at 19:39 0 comments

    We've stuck a little bit with the average current regulation when our supply goes to the CC (constant current) state. The LM51772 has an internal DAC for Iset, or we can set a resistor between ILIM and GND. Our goal is to substitute Riset with our scheme so that we can set it with the DAC of the MCU:

    At first, it was ringing a lot, but after the wires shortened, it became much stable and we can control the current! That was because our auxiliary scheme had become a part of the control loop. In the prototype, we should be very careful with it.

    Here you can see gorgeous soldering PLS to PLS. Anyway really small current goes about 7cm from the chip to our loop...

    Let's get to the result:

    We can control the current from 40mA to 3.4A with a step of 0.7mA. 

    - Shunt should be bigger, not 10 but 20mOhm

    - Add filters to ILIM to adjust the speed and stability of the loop

    - Be careful with routing in the prototype due to high sensitivity and tiny currents.

    - DAC signal should be divided

  • Design Update: PowerBlock-323 Housing

    Andrew10/03/2025 at 01:57 0 comments

    Our team at Encadra has joined the development of PowerBlock-323.

    We are contributing our experience in product design and enclosure engineering to help move the project forward.

    As part of this work, we are now evaluating two enclosure directions: one made from bent sheet metal and another from solid aluminum. Both versions share the same functional layout, but they differ in manufacturing methods and trade-offs.

    We will continue to share design updates as the project evolves. Next topics may include thermal aspects, assembly approach, and finishing options. Feedback from the community is very welcome.

View all 15 project logs

Enjoy this project?

Share

Discussions

Bharbour wrote 09/22/2025 at 17:52 point

Interesting idea! Do you have any BOM cost goals or thoughts yet?

  Are you sure? yes | no

Ioan Larionov wrote 09/23/2025 at 06:51 point

Thank you! We don't know for sure yet, let's see the first batch quantity.

  Are you sure? yes | no

Similar Projects

Does this project spark your interest?

Become a member to follow this project and never miss any updates