Development and repair of switching mode power supplies , battery charger, motor drives and many others required use of oscilloscope with isolated channels. Unfortunately those  oscilloscopes are relatedly expensive even if you buy an used one like TPS2xxx series. The TPS2012 can be found on ebay for about $1500 and having on mind that this is only 2 channel 100 Mhz bandwidth scope the purchase make no sense except for people that need scope with isolated channels on daily base but can't afford a new one.
For people like me that do not use it on daily basis, and already have other scopes, buying such a scope is not the best option, and here the Tek A6909 come in to the game. This is relatively old Sony-Tektronix isolation amplifier produces in two configurations, A6909 - 2 channels and A6907 - 4 channels.
The amplifier is well designed and fully repairable, no custom or potted parts with available service and repair manual /can be found on the link/
Those amplifier has two issues, first is the non standard probe connection, the same used in the old portable Tek 2xx scopes. Find such a probe is a challenge, and the easiest solution is to reworked it for a standard 10x probe.  The second problem is the leaking electrolytes. I both 2 A6909 amplifiers, and all 4 channels had this issue, so most probably you will need to replace them even the unit seems working.
The probe connection rework is relatively simple task. On the 2 channel model both input and output BNC connectors can be installed on the front panel, on the 4 channel model the output connectors should be kept on the back of the unit.
When replacing the connectors keep on mind that this is isolation amplifier, so you should provide a reliable isolation from the enclosure, and between the channels themselves.  Fortunately the front panel is design around isolation, and the rework is very simple. There is small half flange around the connector that need to be removed to allow proper installation of a standard BNC connector.
I used a Molex 0731315013 insulated BNC connector soldered to a small prototype board. The purpose of the board is not so much to take the resistor and the trimer but to provide a reliable attachment for the coaxial cable. Keep on mind that the original cable is with very thin central wire that can be broken very easy.
Since the amplifier impedance is 333.3 kOhm the original probes are with 9.66 Mohm internal resistance and for a proper operation a 666 kOhm resistor need to be add in series to the probe. Hawing on mind that a full calibration will be required anyway the value of the resistor is not so important until is in available for the adjustment range. In my first version i used 1 Mohm resistor, and planing to change it with a selected 665kOhm 1%. The 1% tolerance giving you +/- 6.6 Kohm and there is a good chance to find easy one that is close to the 666.66 kOhm required. A parallel capacitor is required to compensate properly the added resistor. With 1 MOhm resistor a 11-12 pF capacitor is OK, but the best option is to add a 5-30 pF trimmer capacitor.
Here are the resistor Vishay  RN55D6653FRE6 and the trimmer capacitor Knowles Voltronics JR300 i did order.
To fix the leaking capacitors you will need the following values :
82 uF /35 V Nichicon UPJ1V820MED1TD, 3 pcs per channel on A10 board
22 uF / 35 V Nichicon UPJ1V220MDD1TD, 2 pcs per channel on A10 board
22 uF / 25 V Nichicon UPW1E220MDD, 1 pcs per channel on A10 board
10 uF / 25 V Nichicon UEP1E100MDD, Non Polar 1 pcs per channel on A10 board
10 uF / 16V Nichicon UPS1C100MDD1TD, 4 pcs for A6909 and 8 pcs for A6907 on A30 board
220 uF / 25 V Nichicon UPW1E221MPD1TD , 6 pcs for A6909 and 12 pcs for A6907 on A30 board
47 uF / 16 V SMD tantalum  Vishay TMCHC1C476MTRF , 2 pcs for channel on A20 board
220 uF / 35V SMD  Nichicon UWT1V220MCL1GB, 2 pcs for channel on A20 board, plus 2pcs on A50 board.
47uF / 35 V Nichicon UPJ1V470MED1TD, 3 pcs on A40 board
The really critical capacitors are those on A10 board where all except the 10 uF non polar was bad. on all units. I do recommend to change at least those on A10 and A20 board. The rest of them probably was OK but i change all of them just to be sure there will be no a capacitor issue in the next 10+ years.
Looks like there are at least two versions of the A20 board, the one descried in the service manual having R123 and R692 trimmers and the one i have in my units with no trimmer resistors and only one trimmer capacitor.
A10 board also looks to have some variations, in one of the units i have the 10 uF non polar capacitors and on the second unit i do not have them.
Both units seems to have some trouble with autocalibration, but this is a minor issue. The only popups of autocalibration is to set the output offset. At first place i am not sure this calibration works properly from the beginning, because it seems performed only on two of the ranges, and based on my observation this is not enough. The result is an DC offset in the output that can be easy compensate from the front panel.
Unfortunately the calibration procedure in the service manual is not well described. There are two version of the service manual with slight difference in the calibration procedure. One do not required probes attached when initial offset check is performed and another required probes with shorted input. It's logical to have a shorted probe for all checks that are performed with 0 V.
In addition a 850 V source is required for one of the steps, but instead a low voltage source can be used if the voltage applied directly to the amplifier input.  If apply directly to the amplifier input a 28.308 V are required. To make it easy is good the amplifier end of the added resistor to be a bit longer that will allow you to attach a  small wire clip.
For the rest of the calibration steps a voltage calibrator is required, but any adjustable power supply plus a 4 1/2 voltmeter will do the job.
My initial idea was do design a custom probes, and probably when have more time i will design a custom probes with better maching and proper insulation. Keep on mind that most of the scope probes are 300 V rated, so even this unit is 850 V rated, using a standard scope probes should be done with care, and in to their specified range.
With this rework a standard 1:100 probe can be used, and this is an option if you measure high voltage signals, but in many case the signal is just few volts but with high potential to the ground /gate drivers etc/ . In those cases 1:100 probe is not a good idea, a 1:10 or even 1:1 probes are used, but they should have the proper insulation.