3.2nm Resolution Capture/ Holmium Oxide Absorbance Spectrum 0001c
A project log for DAV5 V2.1 Spectrometer
The ZEN of simplicity Making the Complex...Seem Simple
david-h-haffner-srDavid H Haffner Sr 12/30/2017 at 09:590 Comments
UPDATE: 12/30/2017 4:25:AM
Raw data;
Png image of Holmium oxide spectrum;
Raw data both reference spectrum and sample spectrum;
Absorbance %;
Transmission %;
This figure is arrived at by averaging the 5 major peaks wavelength values; 1) 413.39
2) 447.9 3) 483.4 4) 534.67 5) 640.18 and inserting them into this equation;
1 X slit width X ln/mm (diffraction grating) = spectral band width
R = wavelength / spectral bandwidth
= Resolution
So, we work this equation for each of our peak wavelengths and average it out to get our spectral resolution.
Example;
1st peak; 413.39nm
1 X .10 X 1540 = 154
R = 413.39/ 154 =2.68
R = 2.68nm
We do this for the rest of our peaks and then average them out to get our final result
R = 3.2nm
